MG » Tue Feb 01, 2005 7:40 pm
Mismatch issues would be if your samples were dissolved in ACN or MeOH, you'd get distorted peaks or split peaks. Since they are in water, that shouldn't be a problem.
And even the peaks of the internal standard (which has the same concentration in each calibrator) get bigger the higher the concentration of the calibrator is.
I missed that the first time. I can think of several possible causes for this.
(1) Your instrument response is drifting upward throughout the run, perhaps due to not being equilibrated long enough, or perhaps there is another problem. To check this, pick a mid-level standard and reinject it several times (5 or more) and see if the response is stable.
(2) A preparation error. To check this, reprepare the curve yourself if someone else prepared it the first time. Also make sure your analyte stock solution isn't contaminated by ISTD.
(3) If your ISTD coelutes with an analyte and is structurally similar, you could be observing signal that appears to come from your ISTD but really comes from the analyte (there are a couple of ways this could happen).
To check #2 or #3, inject a mixture of your analytes that has not been spiked with ISTD. If you still see signal for ISTD, you know that either your analyte stock solution is contaminated with ISTD, or your signal for ISTD is nonspecific and coming from one of your analytes.