I was involved in some trace analyses in which the peak I am looking at is small and close to others much larger. However, I was able to see traces reasonably well.
Suddenly I had a septum leak and I replaced it. However, I did not realise the leak until the second injection. During the first injection after the leak I saw no peaks, while the temperature program went on.
In the second injection, I saw very few peaks, with a lot of tailing and shifted retention times, which led me to think immediately about the leak.
There was no head pressure in the gauge.
Nevertheless, once the septum was replaced and the pressure back on, the column had lost resolution. The small peaks were not as defined as before and all the others had widened enough to cover a considerable proportion of my target trace peak.
At the beginning I thought it was just a superposition of my pattern and septum bleed what caused the peaks to increase its width and overlap my trace peak. This is not the case, as the blank gives a perfectly fine baseline and I have cleaned and conditioned the septum as well to rule out such possibility.
I suspect that, during the first injection after the leak, the temperature program went on and there was no gas going through. This could have caused the column to degrade and, therefore, lose resolution.
What are your thoughts about this?
Regards