Make a five point linearity study for your GLP test method validation. typically, target, 80% target, 90%, 11%, and 120% can be used.
After that, make your standard and target samples of concentrations near that 100% target concentration, that's what we do for both GLP and for cGMP.
You'll also need to do the typical precision, accuracy, robustness, etc. Here we also need to do a protocol before starting (get it signed) and a validation report, our QA department demands that.
Running several samples in both labs, in parallel, does not document consistency either. Because if these several samples represent 100 % concentration, or content or whatever, then there is still the possibility of ending with the same result no matter what the real concentration is.
At the companies I’ve worked for that number used to be 2% and I don’t recall too many difficulties living up to it. Maybe because of the calibration procedures?tech transfer that i have been charge of required 4% difference top and it is not as easy sometime to get
Lows of physics are the same, but conditions certainly not. It could be temperature, it could be flow cell volume or even flow cell light path, it could be a polluted chemical used to prepare the mobile phase and yes different hardware makes and even models render different detection behaviour. All these things and more, can cause, and typically will cause different linear range and especially standard curve slope. Your example illustrates the issue quite well and there are many more examples of the kind.thanks to physics behaving the same all over the world a method is yes linear on different systems from different vendors unless there are some very excentric differences in the specs of the modules
Actually I was talking about 3 point (level) calibration and not only during the validation, but every time a calibration is performed. The minimum 3 levels are needed in order to create a cure with a relevant slope. During the validation requires are typically 3 curves, 5 levels.you are correct to speak of a set of 3 calibration curves, that is how we do it during validation.
The effect for the 100% zone maybe minimal, but that’s what you don’t know – i.e. whether or not the samples (unknowns) contain 100 % the same amount as the standard, hence the term “unknownsâ€right again the slope never goes through 0 but it not more then Y intercept then the effect for 100% zone values are minimal and so we can yes go to single point, because we are in a linear zone, thanks to physics