Let me introduce my problem to you,
I`m trying to set up a routine method to detect metanephrine and normethanephrine in plasma via LC-MS-MS. We are using a triple quad 6410 from agilent technologies.
The method consist of 3 parts.
First the manuel sample preparation via SPE.
For the extraction i`m using waters MCX columns 1 CC because in literature they are always having the highest recoverys , about 70 - 80 %.
The problem is i`m only getting 20 %.
The protocol for the SPE looks like this
condition : 1 mL MeOH
equilibrate: 1 mL Water
Load: 1 mL sample ( just pure MN in water, i`m first testing this because i want to exclude protein problems from the plasma)
Washstep1 : 1 mL HCL 0.1 N
Wasstep 2 : 1 mL MeOH
Elute: 1 mL NH4OH 5% in MeOH
After the Elution it gets dryed @ 45 C and under a low stream of N2. afterwards it gets reconstituted in 100 uL 95% (99.9 % ACN + 0.1% F.A.) and 5 % (ammoniumformate 100 mM). 75 uL gets injected into the HPLC. I need to inject so much else i don`t get my LOD.
For the HPLC separation
I`m using a Atlantis HILIC column from Waters.
Gradient with 2 mobile fases
A: 99.9% ACN + 0.1% F.A.
B 100 mM Ammoniumformate
Gradient starting At 95 % A and 5 % B
In time going to 80 % A and 20 % B
And then back to 95 % A and 5 % B
Separation is always good and no problems occur.
For the MS
We are measuring in MRM mode via elektrospray ionisation for 3 ions of each molecule.
The only thing that worries me is the high level of ACN. Somebody told me it is not good for the ionisation? can somebody explain this for me ?
So in the end i`m only getting 20% recoverys. already took the loadstep washsteps and analysed them , no losses in here...
Hope you experts can help me on this one