by
unmgvar » Mon Sep 06, 2010 8:05 pm
Dear Karl
HPLC systems are periodically validated to show that they function correctly,
it is done using PV/PQ protocols that use the kits shown in the website i gave.
when all is done the system passes the test, but more important those tests and kits are capable of showing if something is wrong and not to specs and thus to know what to fix,
in most cases that i know of those procedures for the HPLC tend not to use a column but a long piece of tubing that creates volume delay and back pressure
a column after all is a variable that could influence the test and like you state could influence results.
application wise it is always good to have a simpler system suitability test that is specific to your specific method and will allow you to check the system in combination with the column and method parameters
then you generally check for
RT
RSD
K'
resolution
tailing
PQ of a system will need of you to test for:
pump- flow and gradient. you could show RT precision using a compound but this does not show flow precision and reproducability
autosampler- you need to show RSD mainly, carry over, you do not need a column again here as well,
carry over test is actually a joke in a PQ, it could happen for several factors, some of them being system wide contamination, or old spare parts like rotor seal and injector seat seal needing replacement (they generally are in a preventing maintenance routine)
column oven, you need a thermometer here only. these days to show a good performance you should test at several points for a long period of time, at least 3 and also show repeatability. that takes a few hours to do
UV- linearity, drift & noise WL accuracy- of those the tricky one for you is the linearity. especially these days with all the new UV with new ranges. the old ones go to 1.5AU, other 2 AU or 2.5AU or 3 AU
or you could simply limit yourself to a range that works for all and limit your methods to it
So in general you do not need a column and you should actually avoid it when you test your HPLC