dual wavelength methods
Discussions about HPLC, CE, TLC, SFC, and other "liquid phase" separation techniques.
8 posts
Page 1 of 1
Does anyone have experience using dual wavelength detection for assay and impurities determination? Have you ever compared the area of an impurity at one wavelength with the area of the parent compound at another wavelength (using a relative response factor)? Obviously, one impurity has poor absorbance at the lambda max of the main analyte. Any commentary would be helpful.
-
- Posts: 580
- Joined: Fri Mar 16, 2007 11:29 pm
-
- Posts: 1408
- Joined: Thu Jul 28, 2005 2:08 pm
Hi Lanon,
As the energy intensity of detector lamps varys with the wavelength, I’m not completely convinced that a Relative Response Factor would be universally valid i.e. independent of system, time etc.
Maybe, if you include a daily check for (and potential correction of) the RRF (could be using the SST sample) you’ll be able to justify such a procedure. Validation of the procedure would/should be mandatory.
Best Regards
As the energy intensity of detector lamps varys with the wavelength, I’m not completely convinced that a Relative Response Factor would be universally valid i.e. independent of system, time etc.
Maybe, if you include a daily check for (and potential correction of) the RRF (could be using the SST sample) you’ll be able to justify such a procedure. Validation of the procedure would/should be mandatory.
Best Regards
Learn Innovate and Share
Dancho Dikov
Dancho Dikov
-
- Posts: 239
- Joined: Tue Nov 09, 2004 3:06 pm
Dual wavelength is o.k. We use it a lot. RRFs have to be determined for each impurity (clean impurity standards are available?). As RRFs are often cut down to two decimal places (pharmacopoiea says: no RRFs applied between 0.8 and 1.2), I wouldn't care about different lamp intensities, however check recovery on different instruments (PDA, VWD, maybe old+new lamp).
Alex
Alex
-
- Posts: 2846
- Joined: Mon Aug 30, 2004 7:17 am
What´s this business about the lamp intensity? HPLC is calibration dependent, and recalibration or checks needs to be done ever so often. So if you use two different wavelength for two different substances one calibrates as always.
(Was "dual wavelengths" used to describe the method in which two chromatograms at two diff. wavelength were recorded before?).
(Was "dual wavelengths" used to describe the method in which two chromatograms at two diff. wavelength were recorded before?).
-
- Posts: 1408
- Joined: Thu Jul 28, 2005 2:08 pm
Hi Hans,
It is a case of area percentage, relative to the main peak. So, no need for genuine calibration here. Never the less, there will be needed some kind of kontrol of the the ratio, as mentioned earlier. And reworded by Alex
Thanks for that.
Best Regards
It is a case of area percentage, relative to the main peak. So, no need for genuine calibration here. Never the less, there will be needed some kind of kontrol of the the ratio, as mentioned earlier. And reworded by Alex
Best Regards
Learn Innovate and Share
Dancho Dikov
Dancho Dikov
-
- Posts: 2846
- Joined: Mon Aug 30, 2004 7:17 am
Call it check or calibration, if you properly set up the analysis the lamp intensity differences at different lambda is compensated like everything else (provided differences are constant).
8 posts
Page 1 of 1
Who is online
In total there are 17 users online :: 0 registered, 0 hidden and 17 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am
Users browsing this forum: No registered users and 17 guests
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am
Users browsing this forum: No registered users and 17 guests
Latest Blog Posts from Separation Science
Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.
Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.
- Follow us on Twitter: @Sep_Science
- Follow us on Linkedin: Separation Science