Optimal peak height for IC quantitation

Discussions about HPLC, CE, TLC, SFC, and other "liquid phase" separation techniques.

2 posts Page 1 of 1
Hi, I'm using an ICS-2000 running Chromeleon 6.8.

I was experimenting with the signal-noise measurements when I thought of a question:

For quantitation purposes, what it the maximum/optimal peak height you want to aim for? Too high and the peak could flatten at the apex; too low and you'll have to high of a signal to noise ratio.

Thanks!

Two possible answers here, depending on how you propose to adjust peak height.

If you are injecting a given amount, and propose to change peak height by changing the display settings, it doesn't matter (unless you doing manual peak measurement from a printout). The display is simply for informational purposes and can go off-scale without affecting the actual data processing.

If you are changing the amount injected, then so long as you are withing the linear range of your method, it should not matter. The linear range must be determined experimentally by injecting standards at various levels (this is normally part of method validation). The scenarios you describe (peak shape distortion at the high and and poor S/N ratio at the low end) represent the limits of the range.
-- Tom Jupille
LC Resources / Separation Science Associates
tjupille@lcresources.com
+ 1 (925) 297-5374
2 posts Page 1 of 1

Who is online

In total there are 56 users online :: 1 registered, 0 hidden and 55 guests (based on users active over the past 5 minutes)
Most users ever online was 5108 on Wed Nov 05, 2025 8:51 pm

Users browsing this forum: Majestic-12 [Bot] and 55 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry