- Posts: 13
- Joined: Tue Jul 27, 2010 1:06 am
I am just looking for your kind help. I am setting up a used waters HPLC 1525 system. When I am checking my system I am unable to get a stable baseline on acquisition bar. I am using normal phase column with hexane as mobile phase. I know I have to equilibrate the column for 20 column volumes or almost 1 hour. But even then baseline starts from 3 digits after decimal pints (AU) i.e. 0.001 AU and start going on up to 5 or 6 digits after decimal (AU) i.e. 0.00005 or 0.000003 AU and I am not getting it stable. It appears like a noisy baseline. I degassed my mobile phase, checked the flow cell and even run the system without the column. There is now air bubble in flow cell but when I by pass the flow cell from detector way by taking the flow cell out, I get stable baseline. Please tell me where is problem? Should I call waters service person. I will be really thankful.