Hi,
I'm running biologigal samples on Resource Q anion exchange column + NaCl gradient from 300mM to 1500mM. Any tricks to clean the column from DNA between injections? My run time is 30 min already and I wouldn't like to increase it anymore.

thanks
Dear MK,

If I recall correctly, the resource Q is a porous strong anion exchanger. NaCl gradients, especially when relatively short, may allow DNA to remain on-column (Carryover). This is especially true for porous ion exchangers (as opposed to pellicular phases, such as the DNAPac columns).
In our hands, longer gradients minimize, but not eliminate this effect. Use of sodium perchlorate will also help minimize this carryover, as will elevated pH (~12) with either eluent. If the column will tolerate it, addition of solvent, such as acetonitrile may also help.

Cheers,
Jim Thayer
Dionex Corporation.
2 posts Page 1 of 1

Who is online

In total there are 15 users online :: 2 registered, 0 hidden and 13 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am

Users browsing this forum: Google [Bot], Semrush [Bot] and 13 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry