msa analysis

Discussions about GC-MS, LC-MS, LC-FTIR, and other "coupled" analytical techniques.

4 posts Page 1 of 1
I am trying to find a technique for analysis low levels of msa in dmf. Have tried ilc but msa elutes with formic and acetic acid on the system I have. The sample contains both these. Anybody got any ideas?
GCguy
gcguy,

These analytes are easily separated via ion exchange. With a suppressor, the eluent ions can be removed and the eluent can then be directly coupled to MS. The IonPac AS11-HC from Dionex is a good choice for this separation.

What sort of column/eluent would do this seperation? I am currently using a metrohm supp 5 with a carbonate/bicarb eluent without achieving a seperation.
gcguy
GCguy
gcguy

As I mentioned, the best option is the IonPac AS11-HC. This column uses a hydroxide eluent. If you are only interested in these three analytes (MSA, formic, acetic) then this can be handled under isocratic conditions with an eluent concentration around 5 mM. Generally, however, this column is used under gradient conditions in order to apply it to a wider range of analytes.
4 posts Page 1 of 1

Who is online

In total there are 411 users online :: 0 registered, 0 hidden and 411 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am

Users browsing this forum: No registered users and 411 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry