Hi Stephen,
This is not uncommon: just ion suppression on your deuterated standard. If you need these highly concentrated standard solutions, there is nothing you can do to avoid this.
But the good news is that it should not be a problem for your quantification: for identical concentrations, the IS response in "real" samples will be as much affected as that in the calibration solution.
Best,
Gaetan
I would phrase that as "the IS response in 'real' samples SHOULD be as much affected as that in the calibration solution."
You don't know until you test it out and validate it.