Discussions about GC-MS, LC-MS, LC-FTIR, and other "coupled" analytical techniques.

3 posts Page 1 of 1
Any thoughts on why a calibrator analyzed, updated in the calibration table and saved, would not return the set concentration when analyzed by the method?
Six analytes all at 150ng/ml. One of the six comes up 149.93 ng/ml.

A calibrator is analyzed. Calibration table is updated and method saved. Reprocess the calibration file using the method. 149.93 ng/ml for one of the analytes. 150 ng/ml for the rest. I have seen also 149.99 ng/ml but not 150 ng/ml.

Thank you for any ideas.

This sounds like it migh be a software issue. You do not say what software you are using or the nature of the calibration (single point curve or multipoint calibration with replacement of a single level...)

Chemstation. Single point.
Spoke with Agilent and was told it was a Chemstation-Microsoft conflict, problem, issue...something. They didn't know.
Reinstall software was the answer I was told.
Thanks
3 posts Page 1 of 1

Who is online

In total there are 19 users online :: 2 registered, 0 hidden and 17 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am

Users browsing this forum: Amazon [Bot], Semrush [Bot] and 17 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry