pressure fluctuation on lc-8A
Discussions about HPLC, CE, TLC, SFC, and other "liquid phase" separation techniques.
7 posts
Page 1 of 1
what kind of pressure fluctuation is normal? does it depends on instrument? if so, could anybody has lc-8A system tell me what kind of pressure fluctuation you have . thanks a lot. i have 10bar fluctuation.......
-
- Posts: 621
- Joined: Tue Nov 27, 2007 1:20 pm
A 10 bar (145 psi) pressure fluctuation is bad on any HPLC system. the only time I've seen any thing that bad was in a (normal phase) method with a very low column head pressure. (solved by placing a 500 psi flow restricor in line)
I'd check the mobile phases are degassed, the pump heads don't contain static bubbles and then sequentially check the solvent path up to the column head.
I'd check the mobile phases are degassed, the pump heads don't contain static bubbles and then sequentially check the solvent path up to the column head.
Good judgment comes from bad experience, and a lot of that comes from bad judgment.
-
- Posts: 214
- Joined: Wed Dec 30, 2009 4:51 am
thanks. my total pressure is 200bar at 6ml/min. is total pressure related to the fluctuation you can tolerate? the more total pressure you have, more fluctuation you will have?
and I think mobile phase, air bubble problem is already eliminated.
mobile phase degas at vacuum with ultrosonic for 30 minutes. and used 30ml/min flow rate to purge for 5minutes.
and I think mobile phase, air bubble problem is already eliminated.
mobile phase degas at vacuum with ultrosonic for 30 minutes. and used 30ml/min flow rate to purge for 5minutes.
A 10 bar (145 psi) pressure fluctuation is bad on any HPLC system. the only time I've seen any thing that bad was in a (normal phase) method with a very low column head pressure. (solved by placing a 500 psi flow restricor in line)
I'd check the mobile phases are degassed, the pump heads don't contain static bubbles and then sequentially check the solvent path up to the column head.
-
- Posts: 1408
- Joined: Thu Jul 28, 2005 2:08 pm
Hi Jiang295,
5% pressure fluctuation is not that bad. I know people that would live with 10 %.
And yes, the relative fluctuation is the relevant one in this situation. Many people work with pressurs in the range of 10 - 20 bars, so in their case it would've been a catastrophy to experiance a 10 bar fluctuation.
Best Regards
5% pressure fluctuation is not that bad. I know people that would live with 10 %.
And yes, the relative fluctuation is the relevant one in this situation. Many people work with pressurs in the range of 10 - 20 bars, so in their case it would've been a catastrophy to experiance a 10 bar fluctuation.
Best Regards
Learn Innovate and Share
Dancho Dikov
Dancho Dikov
-
- Posts: 621
- Joined: Tue Nov 27, 2007 1:20 pm
In my experience, large pressure fluctuations are encountered when the operating pressure within the HPLC is low. The "back pressure" provided by the column is insufficient to "bed in" the pump seals or to correctly seat check vaves resulting in leakage around seals and through the check valves.Is total pressure related to the fluctuation you can tolerate? the more total pressure you have, more fluctuation you will have?
In standard RP-HPLC, if the operating pressure is ~2000 psi I would expect a variation of < ±1% on a well functioning system.
Good judgment comes from bad experience, and a lot of that comes from bad judgment.
-
- Posts: 1408
- Joined: Thu Jul 28, 2005 2:08 pm
Lucky you!
What brands of HPLC equipment do you have? Less than 1% fluctuation is a remarkable achievement – I’ll have to admit.
Best Regards
What brands of HPLC equipment do you have? Less than 1% fluctuation is a remarkable achievement – I’ll have to admit.
Best Regards
Learn Innovate and Share
Dancho Dikov
Dancho Dikov
-
- Posts: 214
- Joined: Wed Dec 30, 2009 4:51 am
thanks a lot for all your reply.
Today, I found that pressure fluctuation probably coming form insufficient solvent feeding from reservoirs.
flow rate is 25ml/min if i disconnect pump inlet. and my desire flow rate is around 18ml/min. Is this flow rate not enough since pump spend more time in delivering solvent than refilling solvent?
Today, I found that pressure fluctuation probably coming form insufficient solvent feeding from reservoirs.
flow rate is 25ml/min if i disconnect pump inlet. and my desire flow rate is around 18ml/min. Is this flow rate not enough since pump spend more time in delivering solvent than refilling solvent?
Hi Jiang295,
5% pressure fluctuation is not that bad. I know people that would live with 10 %.
And yes, the relative fluctuation is the relevant one in this situation. Many people work with pressurs in the range of 10 - 20 bars, so in their case it would've been a catastrophy to experiance a 10 bar fluctuation.
Best Regards
7 posts
Page 1 of 1
Who is online
In total there are 12 users online :: 2 registered, 0 hidden and 10 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am
Users browsing this forum: Google [Bot], Semrush [Bot] and 10 guests
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am
Users browsing this forum: Google [Bot], Semrush [Bot] and 10 guests
Latest Blog Posts from Separation Science
Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.
Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.
- Follow us on Twitter: @Sep_Science
- Follow us on Linkedin: Separation Science