Calibration with Me-C23:0 in milk matrix

Discussions about GC and other "gas phase" separation techniques.

5 posts Page 1 of 1
I wonder if a dilution series and line equation of (Me-C 23:0) Methyl Tricosanoic Acid as internal standard(I.S) and using it for quantitation of
C 4:0 till C 20:0 in milk is fairly free from errors in terms of accurate recovery calculation.
I wonder if a dilution series and line equation of (Me-C 23:0) Methyl Tricosanoic Acid as internal standard(I.S) and using it for quantitation of
C 4:0 till C 20:0 in milk is fairly free from errors in terms of accurate recovery calculation.
http://www.chem.agilent.com/en-US/Searc ... &liid=1322
you can refer Agilent Application follow as above link. I think that analysis of FAMEs is so easy. If any question, you can contact me via email
Agilent CE

CE's optimism notwithstanding I would expect some inlet discimination over that molecular weight range. As long as this is consistent from standards to samples it might not necessarily compromise the results.

Peter
Peter Apps

I would prefer in ISTD in the middle of your range of analytes - such as C 19:0 or C 17:0. It would react similar to your unknowns.

That said, FAMES are realtively simple and the methodology is pretty straightforward.
I wonder if a dilution series and line equation of (Me-C 23:0) Methyl Tricosanoic Acid as internal standard(I.S) and using it for quantitation of
C 4:0 till C 20:0 in milk is fairly free from errors in terms of accurate recovery calculation.
From analytical point of view C23:0 could not be so good for splitless injection due to less volatility respect to low boiling methyl esters and discrimination can be present. Anyway you have not many more choice with milk fat. C13:0 is present at 0.15%, C15:0 at 1.2 %, C17:0 at 0.66%, C19 and C21 are easily overlapped with C18 unsaturated and anyway present at least to 0,5%. C23:0 is present into the milk fat between 0,01 to 0,02 percent.
As alternative you can add after methylation another volatile substance that elute into middle part of chromatogram such as a ethyl ester of C12 or 14. This is not a true ISTD but can correct the GC injection.
5 posts Page 1 of 1

Who is online

In total there is 1 user online :: 0 registered, 0 hidden and 1 guest (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am

Users browsing this forum: No registered users and 1 guest

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry