Temperature range for ACE 300 C4 phase

Discussions about HPLC, CE, TLC, SFC, and other "liquid phase" separation techniques.

4 posts Page 1 of 1
I am currently testing the 3 µm ACE 300 C4 RP. Unfortunately I wasn't able to find information about thermal stability for this particular phase. Can anyone comment on this? (I would like to run it routinely at 60 °C.)
Thanks in advance!

I've used this particular column at high temperature (55° C) once.
Just like all other silica based column materials it's stable at least up to 60° C. I wouldn't hesitate to apply even higher temperatures (e.g. 70° C).

Best Regards
Learn Innovate and Share

Dancho Dikov

I would recommend a temperature of 40C or less at pH 7-9, and would recommend a similar temperature range < 40C if working below pH 3 for best column lifetime.
Thomas J. Waeghe
MAC-MOD Analytical, Inc.
Chadds Ford, PA 19317
800-441-7508
twaeghe@mac-mod.com
www.mac-mod.com

OK, then I will stick to 40 °C. (Usually we analyze our samples at pH ~2.5)
Thanks!
4 posts Page 1 of 1

Who is online

In total there are 52 users online :: 1 registered, 0 hidden and 51 guests (based on users active over the past 5 minutes)
Most users ever online was 5108 on Wed Nov 05, 2025 8:51 pm

Users browsing this forum: Baidu [Spider] and 51 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry