Phycal » Mon Jun 28, 2010 11:55 am
													
 
						
					 
					
						The reason I use a lower temperature is due to the fact that most columns I use for RI separations (sugar and organic acids) are heated higher than the maximum temp allowed for the RI detector. It often is easier to have the detector set lower and therefore the temperature fluctuations are less. This has been true for every manufacturer of LC equipment I have used (Waters, Shimazdu, Thermo, Agilent).
When I have been able to use I higher temperature, I often found out you have to insulate the tubing between the detector and the column since a fluctuation in the amount of heat needed causes the baseline to rise or fall.
That is classic pump noise or pulse dampener noise. If an air bubble is trapped in the system somewhere, you will see this type of noise (although the noise is usually has a faster periodicity to it).