Capillary column connection with mass spectrometer

Discussions about GC-MS, LC-MS, LC-FTIR, and other "coupled" analytical techniques.

3 posts Page 1 of 1
hello to all forum
I need some help
in our laboratory we have a gas chromatograph (Varian CP-3800) coupled to a mass spectrometer (Varian 1200)
Yesterday we decided to change the capillary column and when we put the new column, do not know the exact length of column that is inserted into the chromatograph injector (mod. 1079) and the transfer line to mass spectrometer
if anyone can help me, I would be eternally grateful
hello to all forum
I need some help
in our laboratory we have a gas chromatograph (Varian CP-3800) coupled to a mass spectrometer (Varian 1200)
Yesterday we decided to change the capillary column and when we put the new column, do not know the exact length of column that is inserted into the chromatograph injector (mod. 1079) and the transfer line to mass spectrometer
if anyone can help me, I would be eternally grateful
Juanma,

I don't have a Varian but normally is simply a matter to adjust the distance of the column (on my systems the column is protuding 1-2 mm from the heated transfer line). If your system doesn't allow a direct visual control use a septum to mark the correct position.

You should have a user manual where the procedure is described.

https://snri.ucmerced.edu/snri/eal/Vari ... 914978.pdf (it's not the same instrument but maybe you can find something useful at page 127 of the document).


bhuvfe

hello bhuvfe

firstly I would like to thank you for the information that you have provided to me
as you well say, the ideal is that had visual access to the termination of the capillary, but the ionization source of spectrometer prevents me.
I saw the link you put in your post and I must say that the information is very good, but the transfer line is different from the one we have installed, thus, distances are not the same
finally inserted until the end of the column and stepped back about 5mm and then tighten the nut.
for now it seems the instrument works well, but I would be under vacuum pressure, but I imagine that when within one day get it.

I repeat, thank you for your help

receives a greeting

Juanma
3 posts Page 1 of 1

Who is online

In total there are 370 users online :: 1 registered, 0 hidden and 369 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am

Users browsing this forum: Semrush [Bot] and 369 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry