strange over pressure problem

Discussions about HPLC, CE, TLC, SFC, and other "liquid phase" separation techniques.

2 posts Page 1 of 1
Hi

I'm having a strange over pressure problem. I'm using a C-18 column, and reverse phase chromatography. I'm trying to detect catecholamines (dopamine in particular), and my mobile phase is NaH2PO4, EDTA, Octanesulfonic acid, and 10% ACN. Here's my problem: When the OSA is included in the mobile phase, I have overpressure problems. When I "clean" the column by running 20 void volumes of the mobile phase w/o the OSA (ion pairing agent) and then run 20 void volumes of h20, 10% MeOH, the pressure goes down. It stays in a normal range (about 2k psi). When I go back to a mobile phase with the octanesulfonic acid, the pressure goes back up to about 3600+ psi.

Any ideas why this is happening? Should I just buy a new column?

I'd prepare that MP by combining your phosphate, solids and water, mixing and filtering that before introducing the ACN.

I suspect you may have some undissolved EDTA or OSA (or common impurities in either) fouling things up.
Thanks,
DR
Image
2 posts Page 1 of 1

Who is online

In total there are 23 users online :: 3 registered, 0 hidden and 20 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am

Users browsing this forum: Ahrefs [Bot], Baidu [Spider], Google [Bot] and 20 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry