Problem of needle plug by septum in gas tight syringe

Discussions about GC and other "gas phase" separation techniques.

4 posts Page 1 of 1
In analyzing gas using gas tight syringe (manual injection), we meet the problem of needle clogging rather frequently (almost every day). It's all septum debris inside the syringe. For micro syringe (liquid) we almost never have such a problem.
Have you got any idea how to overcome this case?

Use a cone-tipped or dome-tipped needle, not the bevelled tip that for some reason is the default on micro-syringes despite its being the worse design for tearing chunks out of the septum.

Peter
Peter Apps

Dome tip with side hole :wink:

Thank you both for the help. But I am still confused why use small syringe (5-10ul) that problem rarely happens and the needle is exactly the same: pointed bevelledly.