May people could accept the reversal peaks appeared between 1 to 5 min, and that is said as the peaks of solvent. While how about the reversal peak after the target peaks? How to explain this phenomenon? I am looking forward to your explaination. Thanks!

You probably mean negative peaks? That's my assumption anyway.
If the stuff causing a negative peak absorbs less energy (my second assumption is that you're utilizing an absorbance detector) than the mobile phase does, at the given wavelength, you'll observe a negative peak, regardles of its retention time.

Best Regards
Learn Innovate and Share

Dancho Dikov

Much more information from you is required to be able to help you.

I know of a IC application for example; where a negative peak at the tail of the main compound is exactly what you look for and try to quantify

What is your application?
3 posts Page 1 of 1

Who is online

In total there are 22 users online :: 3 registered, 0 hidden and 19 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am

Users browsing this forum: Amazon [Bot], Google [Bot], Semrush [Bot] and 19 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry