how to change MS display range in Chemstation

Discussions about GC-MS, LC-MS, LC-FTIR, and other "coupled" analytical techniques.

2 posts Page 1 of 1
Hi,

Does anyone know how to extend MS spectrum display range in Chemstation?

My spectrum showed up to 117 but I am sure there is a molecular ion at 132. Due to the chemstation software setting of display 1-21 ions, I am not able to see anything beyond 120.

Hope someone here has a handle to work around this.
Thanks in advance,
Excel

it is due to your scan range
in the instrument menu go ms sim/scan parameters
click on scan parameters
set your range
2 posts Page 1 of 1