I want to install a Sedex 85LT ELSD on a HPLC system with HP1050 IEEE/GPIB connected components, and run on a Chemstation with version A.08(?). Is there a way to direct the ELSD signal into the Chemstation?
-
- Posts: 3594
- Joined: Mon Aug 30, 2004 11:43 pm
Yes. You need either an internal or external 35900 interface (an analog-to-digital converter).
-
- Posts: 212
- Joined: Fri Jul 20, 2007 7:51 pm
Sally: what Analytes r you trying to quantify with your evaporative light scattering detector?
Jumpshooter
-
- Posts: 154
- Joined: Thu Jan 20, 2005 11:27 pm
We have already developed and validated HPLC-ELSD methodologies for Gentamicin and Neomycin. We currently run HPLC-ELSD for these analytes on another HPLC system; but, due to work load and other considerations, we want to move the ELSD to the HP1050 / Chemstation HPLC system.
-
- Posts: 50
- Joined: Wed May 06, 2009 12:08 am
Hi, Sallybeetle -Yes. You need either an internal or external 35900 interface (an analog-to-digital converter).
You'll also have to use the external contact closure cable that (on the HP1100, anyway) attaches to the back of the pump. This is to signal the instrument to start/stop runs and to autozero. In my experience, the contact closure cable didn't connect easily to the ELSD - you have to get the correct cables and connectors - but it works well after doing it.
Also, if you want to shut off gas after the run, you can get a Gas Mizer - a nifty little item sold by (I believe) Ticoscen. There are lots of posts on here from Tim at Ticoscen. He'll help you out.
H_H
-
- Posts: 6
- Joined: Tue Mar 17, 2009 3:05 pm
Hi Sallybeetle,
rgds.
S.Scala
would you mind to share some details about the methods for Gentamicin and Neomycin?We have already developed and validated HPLC-ELSD methodologies for Gentamicin and Neomycin.
rgds.
S.Scala
-
- Posts: 3594
- Joined: Mon Aug 30, 2004 11:43 pm
1050 pump and autosampler modules each have contact closures (believe both as contact #2).
7 posts
Page 1 of 1
Who is online
In total there are 27 users online :: 1 registered, 0 hidden and 26 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am
Users browsing this forum: Google [Bot] and 26 guests
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am
Users browsing this forum: Google [Bot] and 26 guests
Latest Blog Posts from Separation Science
Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.
Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.
- Follow us on Twitter: @Sep_Science
- Follow us on Linkedin: Separation Science