the final eluent I am using is 2 bed volume of 2M methylamine in methanol,
I also used o-AMINOPHENOL as internal standard, but it did not elute out either.
did you see any problem?
Advertisement
Discussions about HPLC, CE, TLC, SFC, and other "liquid phase" separation techniques.
Why 100% methanol in ion exchange chromatography?
What is pKa of your analyte?
Best Regards
Why 100% methanol in ion exchange chromatography?
What is pKa of your analyte?
Best Regards
Not at all - in ion exchange chrom. anyway! I'm afraid you're mixing IEx. and RP in a non controllable manner.is NH3 in methanol very normal for final eluent?
Not at all - in ion exchange chrom. anyway! I'm afraid you're mixing IEx. and RP in a non controllable manner.is NH3 in methanol very normal for final eluent?
If you intend to utilize the IEx. technique, you'll need a reasonably high salt concentration in the eluting solvent and the most of it must consist of water, rather than an organic solvent like methanol.
Best Regards
The first thing I would call wrong (not the equation) is the idea of using 100 % ethanol as a solvent.
I still do not comprehend the rationale for substituting methanol for water.
As for the equation; nothing wrong provided the solvent was water. You see I’m far from convinced that the ionisation either of the analyte or the stationary phase itself is anywhere near that complete.
So, if you can find a rationale for utilizing 100 % methanol instead of water, the set-up will seam much more convincing.
Best Regards
The first thing I would call wrong (not the equation) is the idea of using 100 % ethanol as a solvent.
I still do not comprehend the rationale for substituting methanol for water.
As for the equation; nothing wrong provided the solvent was water. You see I’m far from convinced that the ionisation either of the analyte or the stationary phase itself is anywhere near that complete.
So, if you can find a rationale for utilizing 100 % methanol instead of water, the set-up will seam much more convincing.
Best Regards
Hi again,
In order to perform an ion exchange chromatography, it is absolutely essential for both the analyte and the stationary phase to be charged (ionized). That’s what IEx is all about. If one of these or both of them isn’t charged you simply don’t perform that kind of chromatography.
So…… more concrete in your case and provided the analyte’s not charged, you either lose it during the wash phase, or it stick to the stationary phase due to hydrophobic or more probably hydrophilic interactions.
Best Regards
P.S. If you feel like sharing some info and/or data that contradict the above, please do so. I’m convinced that the most of the knowledge that I’m going to acquire during my life still lies ahead of me.