The method is an LC-MS MRM method screening over 60 pesticides in a 8 min gredient. LC from Agilent,and QQQ from AB Sciex.
Wired things happend after last two PMs. Right after the PM in last fall, I lost sensitivety of all my compounds from 50 to 80%. No culprit found, and after one or two month, it went back by itself to normal. No tuning or calibration was done during this period.
Just last week, right after this PM, we lost signal (total loss) for about 45 compounds, however, the detectable ones have no RT shift or sensitivity loss. For both group of compounds, their RT and Q3 m/z are all over the range (1 to 7 min, and 60 to 400m/z).
Eevery PM parameter passed the requirment; Re-calibrated the mass; LC was proven to have no problem by infusion and other checks; ESI probe was good but replaced with new one; method was not corrupted,proved by re-writing, instrument was "baked" over weekend, no improvement; manual optimziation for some lost compounds found no signal; no problem with standards......
Problem is still unsolved by now
What do you catch? Thanks and have fun