which factors effect relative retention time
Discussions about GC and other "gas phase" separation techniques.
3 posts
Page 1 of 1
In order to locate some interested solvents, I performed the method described in ChP 2010 version, non-polar column, maintain at 40℃ for 8 min, then with 8℃/min to 120℃, maintain 10min. In this method, I use a DB-1 column with different ID and film described in ChP. But the result is disappointed. It is hard to locate the solvents by their PARTs. PART=(tR-t0 ) /( t’R-to), where to is the retention time of Methane, t’R is the retention time of Butanone, tR is the retention time of interested solvent. Would you kind to explain which factors effect PART and how they affect it . Thanks
-
- Posts: 1680
- Joined: Sat Aug 23, 2008 12:04 am
When you change column dimensions, compounds will elute more quickly from a column with a narrower diameter and from a column with a thinner stationary phase.
You must also take into consideration the linear velocity of the mobile phase. If you have operated the column at the same pressures as used for a column with different dimensions, the peak shape may change, degrading resolution and making it difficult to find some peaks.
The temperature ramp is set to help resolve peaks at the rates at which they elute from the original column. If you have changed column diameter, expect to need to change the temperature profile as well.
You must also take into consideration the linear velocity of the mobile phase. If you have operated the column at the same pressures as used for a column with different dimensions, the peak shape may change, degrading resolution and making it difficult to find some peaks.
The temperature ramp is set to help resolve peaks at the rates at which they elute from the original column. If you have changed column diameter, expect to need to change the temperature profile as well.
-
- Posts: 5433
- Joined: Thu Oct 13, 2005 2:29 pm
The problem might lie with the retention time of methane - how are you measuring this ?
Peter
Peter
Peter Apps
3 posts
Page 1 of 1
Who is online
In total there are 56 users online :: 2 registered, 0 hidden and 54 guests (based on users active over the past 5 minutes)
Most users ever online was 5108 on Wed Nov 05, 2025 8:51 pm
Users browsing this forum: Baidu [Spider], Semrush [Bot] and 54 guests
Most users ever online was 5108 on Wed Nov 05, 2025 8:51 pm
Users browsing this forum: Baidu [Spider], Semrush [Bot] and 54 guests
Latest Blog Posts from Separation Science
Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.
Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.
- Follow us on Twitter: @Sep_Science
- Follow us on Linkedin: Separation Science