AICMM » Thu May 13, 2010 1:21 pm
CE Instruments,
An new HID for $2500, if you have an old FID electrometer hanging around does not seem like too much money if you compare it to the headache of trying to measure hydrogen at the same time as other fixed gases. It is really cost effective if you are considering two TCD's, one for hydrogen and one for everything else. Then it will beat that price. Especially after you add in a second valve and second column for the second gas channel.
Yes you have to be leak free but you don't have to be pristine by any shake of the imagination. I use copper tubing, and standard flow fixtures from a 5890 or 6890 and I don't have any problems on mine. You also don't need more than five 9's helium and I don't use a getter either. Besides, a system that is leak proof enough for a TCD could very easily be not leak proof enough considering the poor sensitivity of the TCD in many applications.
If you are looking at a couple of hundred ppm to percent levels, which a lot of applications do, then the extra noise of an HID doesn't matter. Not that it is THAT much noisier (about an order of magnitude) if you are leak free. I have looked at 0.5 to 1% of hydrogen on a TCD and you can barely see it. That's a great peak on my HID. I have also looked at 20 ppm with a 30 uL loop if you need to do low levels.
The only significant issue with an HID in this application is working near 100%. Then your peak will be distorted and non-linear compared to a TCD but a lot of people don't care about the near 100% peak.
Back off my soap box...
Don't get the wrong impression. I still believe in the TCD as a detector for certain applications, especially where you cannot use helium. However, the use of TCD for measuring hydrogen and.... is not one of them in my opinion.
Best regards.
P.S., I have run my HID out of helium a couple of times (typically over a weekend) and once I restore the helium flow to it on Monday, it starts right back up. Yeah you have to bake it, mostly to get rid of water (water, water everywhere....) but the detector itself will be fine.