Strange chromatogram due to problem with instrument

Discussions about GC and other "gas phase" separation techniques.

10 posts Page 1 of 1
This is the chromatogram when I monitor my GC system. This is a capillary column with FID detection. An analyst replaced the liner, septum ferrules and then installed her column. The only signal, at first, was just baseline noise, even though there were many injections. Today the column was reinstalled with 10 cm of length from each side of the column cut off. Now, at 50 C I see the following.

Image

Does anyone know whats going on? For some of you more experienced Analysts this might be a no-brain'er. Thanks for any help.
MestizoJoe
Analytical Chemist and Adventurer
Venture Industries
Spider-Skull Island

MestizoJoe: there is definitely something going on with your Detector in that it is giving a carbon-burn signal according to a systematic regular cycle which almost resembles an oscillation pattern.
Please provide us with the run parameters and the identity of the compound being analyzed.
Jumpshooter

have you done any maintenance on your detector recently?

I don't suppose your FID has automatic re-ignition? Your pattern looks like the flame keeps getting ignited then quenched, then re-ignited.

If so I'd try removing the detector, cleaning the area out and sonicaing the torch.

jh1's suggestion is a good one, I have also seen something like this when using a gas generator to supply the make-up nitrogen.

This might also be due to a problem with detector or carrier gas controllers - do you see any fluctuations in pressures and flows ?

Peter
Peter Apps

I think JH1 has nailed the issue, it remains to be seen what is causing the re-ignition.

Leaks, flows out of control, deposits in the flame tip, are all possibilities.

Time for someone who is experienced to fix the darn thing.

Do you have anyone like that?

Rodney George
consultant

Yes! Thank you JH1. The flame did repeatedly reignite. The problem was solved when we cleaned the jet which was clogged. I'm not sure if that was the primary issue but once we put the detector back together everything was fine.

Thanks guys.
MestizoJoe
Analytical Chemist and Adventurer
Venture Industries
Spider-Skull Island

I am not experienced but I fixed it anyway because I am a nice guy.
MestizoJoe
Analytical Chemist and Adventurer
Venture Industries
Spider-Skull Island

You were inexperienced but looks like you are getting more and more of it. Congratulations on the fix.

I hope they appreciate you.

Rodney George

I know you fixed the problem already, but you should also make sure you detector offset isn't set too high.
10 posts Page 1 of 1

Who is online

In total there are 2 users online :: 1 registered, 0 hidden and 1 guest (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am

Users browsing this forum: Nadee and 1 guest

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry