I'm having troubble with one of my GC methods. The GC sequence looks something like this:
2 wash
4 standards
2 wash
10 unknowns
2 standards
2 wash
10 unknowns
repeating 3 last rows depending on amount of samples to be analyzed.
We have a minimum requirement for the area of the standard (which is 2,8).
The 4 first standards are usually around 4 in area.
The problem is then the first of the 2 next standards: 1,7!
Second of the 2 will typically be 3,3.
The same will occur on the next set of standards in the sequence.
It seems the analyte gets absorbed in the GC somewhere. I change syringe, liners, septum and even columns pretty often to keep the GC in shape cause of this. Usually either conditioning the inlet on high temperature, changing liner or changing (or cutting) the column helps when the area starts going downhill. But this time nothing works!
I would be thankful for any advice!