I recently installed a new column in our gc/ms agilent.
The column is rtx-1ms (30m, 0.32 mm, 0.5 um).
The oven program is 100C for 1 min then 20C/min to 300C for 7 min (total run 18 min)
When it is reach around 10 min the baseline starts to bleed!! although the separation is occured but staionary phase ions always appearing.

I changed the septum, liner, reinstall the column at injector port but no sucssedense.

any hints?

First of all: all columns bleed some, some just more than others. If you magnify enough, you will always see some bleed. One tries to have the analyte signal considerably more than the noise or bleed. What is your split ratio and what are your concentrations? Are you getting enough analyte on the column?

If your carrier purification system is not removing oxygen to a low enough level you could be destroying your column.

That would be my guess.

Rodney George
consultant

che313,

You have a leak detector already installed. What is your 28 and 32 doing, how big are they? Any other odd ions at idle?

Best regards.
4 posts Page 1 of 1

Who is online

In total there are 32 users online :: 0 registered, 0 hidden and 32 guests (based on users active over the past 5 minutes)
Most users ever online was 5108 on Wed Nov 05, 2025 8:51 pm

Users browsing this forum: No registered users and 32 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry