krickos » Fri Mar 26, 2010 6:28 am
The main thing is that empty vials have the ethanol peak.
As an update: Since we are using an Agilent HS, we found on the website that we can 'Steam Clean' the lines. We'll try this also today.
Well the Agilent config with your temp settings in loop/valve, t-line and potetially injektor can very much be the reason you have blank issues with ethanol. This phenomena I/we have seen plenty of times.
Bp of DMSO is what 189°C? with your settings you will over time contaminate sample loop/valve with DMSO residues, polar solvents like ethanol will attach to it, while less polar solvents/low boilers such as acetone and toluene is less prone to attach to contaminants.
Steam cleaning will likely help to clean out loop/valve in worst case loop has to be manually cleaned.
So a basic pointers with regard to Agilent headspace instruments:
ALWAYS, set temperatures in valve/loop, transfer line, injector above the boiling point of the higest boilier which includes the sample solvent. By doing this you will have considerble less downtime in long run.
ohh a second comment: Unless you really need it, avoid that low split ratio of 1:1 mÃght cause you even more troubles, try to stay at 1:5 or higher.
If you need trace analysis and really want to do like splitless or split ratio 1:1 you need to go for a high id column like 0,53mm id with higher flow rates in order to get decent peak shapes on to column. You can search Agilent homepage for headspace and splitless for more information.