I am assigned to verified O2 limit test method in NF N2 monograph as per USP <1226>. I am wondering to what extent I shall perform the method verification. I think system suitability is a must, how about other parameters, like limit of quantification (LOD), precision, accuracy?
The test method is showed below:
Sample: Nitrogen Sample
Standard: 1.0% Oxygen in nitrogen certified standard
Chromatographic system
Mode: GC
Detector: Thermal conductivity
Column: 2.1-mm × 2.4-m, packed with 80/100 mesh S3 packing
Carrier gas: Helium (99.99%)
Flow rate: 30 mL/min
Column temperature: Thermostatically controlled at 70 degree C
System suitability
Samples: Standard and Sample
System suitability requirements
Resolution: NLT 2.0 between nitrogen and oxygen, Standard
Retention time: The oxygen peak response from the Sample exhibits a retention time corresponding to that from the Standard.
Analysis
Samples: Standard and Sample
Introduce the Samples separately into the gas chromatograph by means of a gas sampling valve.
Acceptance criteria: NMT 1.0% of oxygen when compared to the peak response of the Standard, indicating NLT 99.0%, by volume, of N2
Thanks in advance,
Terry