I am Juan Manuel, at this moment I try to do a calibration line of silybin. My problem is: If I dissolve with methanol the silybin (20 mg/ 100 ml) I have four peaks, but If I dissolve with methanol/water (50%/50%) (20mg/100ml) I have two peaks. The sylibin have two diastereoisomers.

Other problem is when I dissolve with methanol (20mg/100ml- > 200 ppm) before I dilute with water 100 ppm (50ml of 200 ppm with 50 ml of water) and others dilute. I have not good calibration.

Why is it happening?

Thanks everybody

As to your first problem, perhaps you are experiencing injection solvent mismatch with the methanol? I have seen injections in excessively strong solvent produce split peaks. It would help to know your conditions (column, mobile phase(s), flow rate, injection volume).

Dear Juan Manuel, a similar problem has been published by Hoffman et al. in J. Chromatogr. 465 (1989) p. 189. They analyzed phenylalanine on a RP HPLC phase with buffer/acetonitrile 92:8. When the sample was dissolved in e.g. water/acetonitrile 1:1 the peak looked very crazy because the sample solvent was much too strong related to the eluent.

Veronika Meyer

Heavy peak broadening, although no double peaks, were observed in vitamin analysis if the sample solvent was too strong, see Naish et al., Chromatographia 20 (1985) p. 335.

Veronika Meyer

As to your first problem, perhaps you are experiencing injection solvent mismatch with the methanol? I have seen injections in excessively strong solvent produce split peaks. It would help to know your conditions (column, mobile phase(s), flow rate, injection volume).
My colum is Kromasil C18 250*4,6, flow 1ml/min, injection volume 20 microlitres and mobile phase is 50% methanol 50% buffer to ph 2,8 phosphate

Dear Juan Manuel, a similar problem has been published by Hoffman et al. in J. Chromatogr. 465 (1989) p. 189. They analyzed phenylalanine on a RP HPLC phase with buffer/acetonitrile 92:8. When the sample was dissolved in e.g. water/acetonitrile 1:1 the peak looked very crazy because the sample solvent was much too strong related to the eluent.

Veronika Meyer
hello Veronika, thanks by your answer. At this moment I don´t acess to this articles if you can send me to "juaant@cartif.es" I was very wondeful
6 posts Page 1 of 1

Who is online

In total there are 20 users online :: 0 registered, 0 hidden and 20 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am

Users browsing this forum: No registered users and 20 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry