hi i had a problem notified many time i.e my waters2487/2489 with empower database HPLC system shows an error "lost prime" what it is and how can i solve this problem?
You have air somewhere in the solvent lines that is causing a pressure drop in your system which empower is saying is a loss of prime.
Make sure your mobile phase(s) are full degassed
If you have a Waters 2695/2795 separations module:
Go through the "Dry Prime" procedure on all your solvent lines (including unused ones. Follow with a "wet prime" on all lines. then reactivate the method and monitor.
If you have an standard HPLC pump thoroughly purge the solvent lines and check the pump heads for possible airlocks