regarding my real time problems araised during my analysis
Discussions about HPLC, CE, TLC, SFC, and other "liquid phase" separation techniques.
4 posts
Page 1 of 1
hi i had a problem notified many time i.e my waters2487/2489 with empower database HPLC system shows an error "lost prime" what it is and how can i solve this problem?
-
- Posts: 621
- Joined: Tue Nov 27, 2007 1:20 pm
You have air somewhere in the solvent lines that is causing a pressure drop in your system which empower is saying is a loss of prime.hi i had a problem notified many time i.e my waters2487/2489 with empower database HPLC system shows an error "lost prime" what it is and how can i solve this problem?
Make sure your mobile phase(s) are full degassed
If you have a Waters 2695/2795 separations module:
Go through the "Dry Prime" procedure on all your solvent lines (including unused ones. Follow with a "wet prime" on all lines. then reactivate the method and monitor.
If you have an standard HPLC pump thoroughly purge the solvent lines and check the pump heads for possible airlocks
Good judgment comes from bad experience, and a lot of that comes from bad judgment.
-
- Posts: 580
- Joined: Fri Mar 16, 2007 11:29 pm
also a stucked check valve will produce this failure
if no solvent comes out of the purge valve (on alliance systems), then remove the check valve and replace it or at least do some thorough cleaning by use of a ultrasonic bath.
If you are mixing methanol and water and the problem persists, a) have your degasser checked and b) do some additional helium sparging of your solvents prior you put them on the system.
if no solvent comes out of the purge valve (on alliance systems), then remove the check valve and replace it or at least do some thorough cleaning by use of a ultrasonic bath.
If you are mixing methanol and water and the problem persists, a) have your degasser checked and b) do some additional helium sparging of your solvents prior you put them on the system.
-
- Posts: 129
- Joined: Thu Oct 04, 2007 7:16 pm
Another possiblity could be a clogged in-line filter. If you have an extra one replace it if not you can clean it by sonication for a temporary solution. Most filter clogs tend to be buffer related so I use a (90/10 water/organic).
4 posts
Page 1 of 1
Who is online
In total there are 34 users online :: 1 registered, 0 hidden and 33 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am
Users browsing this forum: Ahrefs [Bot] and 33 guests
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am
Users browsing this forum: Ahrefs [Bot] and 33 guests
Latest Blog Posts from Separation Science
Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.
Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.
- Follow us on Twitter: @Sep_Science
- Follow us on Linkedin: Separation Science