I have a general question: for clinical samples measured by LC/MS/MS, it is required to have internal standard? Thanks for any input.

As far as I know, internal standards are not required so long as you can demonstrate acceptable accuracy and precision when you validate the method.
-- Tom Jupille
LC Resources / Separation Science Associates
tjupille@lcresources.com
+ 1 (925) 297-5374

http://www.fda.gov/downloads/Drugs/Guid ... 070107.pdf

The FDA guidance for industry does not mention the use of IS in assays at all. However, in my experience (20+ years), I have not seen a bioanalytical method for an API that has not been an IS method .

Tom is quite correct in saying that if you can demonstrate acceptable acuracy and precision you don't have to use an IS. However, given the low levels being analyzed and the often complicated extraction procedures associated in exracting the APIs from a biological matrix, an IS method is advisable.
Good judgment comes from bad experience, and a lot of that comes from bad judgment.
3 posts Page 1 of 1

Who is online

In total there are 35 users online :: 0 registered, 0 hidden and 35 guests (based on users active over the past 5 minutes)
Most users ever online was 5108 on Wed Nov 05, 2025 8:51 pm

Users browsing this forum: No registered users and 35 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry