How to flush clogged Valco injector?
Posted: Tue Mar 09, 2010 1:46 am
by Jiri Urban
Hello all,
Does anyone of you has experiences with clogged Valco CHEMINERT (4 nl) injector? Is there any solvent which "eat" all the clogged stuff and do not "touch" the CHEMINERT stator? I was working mainly with uracil and alkybenzenes in reversed-phase mode.
Thanks in advance
Posted: Tue Mar 09, 2010 5:41 pm
by tom jupille
If it's not too badly clogged, you can try running the flow through the valve in reverse (connect the pump where the column used to be). For uracil & alkylbenzenes, any reasonable mobile phase should work (the trick is to use the pump pressure to blow the plug out the short way).
Posted: Tue Mar 09, 2010 6:17 pm
by Jiri Urban
Thanks Tom, I have already tried it, but probably with the lower pressure than necessary. I had small success with the ultrasonication overnight in (ehm) DMSO, so maybe the combination of both will help. Thanks.
Posted: Thu Mar 11, 2010 11:16 am
by Peter Apps
Combining pressure with ultrasound should do the trick.
Posted: Thu Mar 11, 2010 4:29 pm
by HW Mueller
I am not familiar with the Valco, but they used to be similar to Rheodyne. If this is the case it should be worthwhile to relate my experience with the Rheodyne: I only recall clogging of the type where some particles (most likely introduced via the syringe) plugged outgoing capillaries (very small inner diameter). Opening one connection after the other revealed which was the culprit (the capillaries with flow to the valve were usually not involved).
Posted: Wed Mar 17, 2010 8:25 pm
by Jiri Urban
Thanks all. It seems, injector is working properly.
Posted: Wed Mar 17, 2010 8:53 pm
by tom jupille
So, what was the "magic trick" ?
Posted: Sat Mar 20, 2010 5:01 pm
by Jiri Urban
I think the combination of ultrasonic bath and DMSO. However, I still cant "see" through the injection port, eventhough it is injecting.