How to flush clogged Valco injector?

Discussions about HPLC, CE, TLC, SFC, and other "liquid phase" separation techniques.

8 posts Page 1 of 1
Hello all,

Does anyone of you has experiences with clogged Valco CHEMINERT (4 nl) injector? Is there any solvent which "eat" all the clogged stuff and do not "touch" the CHEMINERT stator? I was working mainly with uracil and alkybenzenes in reversed-phase mode.

Thanks in advance
HPLC 2017 in Prague, http://hplc2017-prague.org/

If it's not too badly clogged, you can try running the flow through the valve in reverse (connect the pump where the column used to be). For uracil & alkylbenzenes, any reasonable mobile phase should work (the trick is to use the pump pressure to blow the plug out the short way).
-- Tom Jupille
LC Resources / Separation Science Associates
tjupille@lcresources.com
+ 1 (925) 297-5374

Thanks Tom, I have already tried it, but probably with the lower pressure than necessary. I had small success with the ultrasonication overnight in (ehm) DMSO, so maybe the combination of both will help. Thanks.
HPLC 2017 in Prague, http://hplc2017-prague.org/

Combining pressure with ultrasound should do the trick.

Peter
Peter Apps

I am not familiar with the Valco, but they used to be similar to Rheodyne. If this is the case it should be worthwhile to relate my experience with the Rheodyne: I only recall clogging of the type where some particles (most likely introduced via the syringe) plugged outgoing capillaries (very small inner diameter). Opening one connection after the other revealed which was the culprit (the capillaries with flow to the valve were usually not involved).

Thanks all. It seems, injector is working properly.
HPLC 2017 in Prague, http://hplc2017-prague.org/

Thanks all. It seems, injector is working properly.
So, what was the "magic trick" ? :wink:
-- Tom Jupille
LC Resources / Separation Science Associates
tjupille@lcresources.com
+ 1 (925) 297-5374

Thanks all. It seems, injector is working properly.
So, what was the "magic trick" ?
I think the combination of ultrasonic bath and DMSO. However, I still cant "see" through the injection port, eventhough it is injecting.
HPLC 2017 in Prague, http://hplc2017-prague.org/
8 posts Page 1 of 1

Who is online

In total there are 16 users online :: 2 registered, 0 hidden and 14 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am

Users browsing this forum: satya12, Semrush [Bot] and 14 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry