Chemstation Problem Batch Review
Got two new Agilent 7890A which I am using with Chemstation B04.01 481. I am a newbie at this.
I just want to export my sequence data [areas] via batch report into an excel file to do some ongoing calculations. I got all the output settings but no areas are shown in the .xls
The problem is the list of compounds in batch mode is empty and i dont now how to fill it.
Can someone please help me?
Batch review will only give you areas of peaks identified in a calibration curve.
Check the Batch Report Settings, where you can specify which parameters are included in the table. It isn't very user friendly, but if you play with it you should be able to figure it out.
Also, exporting area/height is OK, but be aware that Batch Review uses it's own scheme for calculations, and it may not match your method.
Merlin K. L. Bicking, Ph.D.