Is there an equation that relates sensitivity with the properties of the GC peak e.g. peak width or peak height?

Don't think it's is a defined chromatographic parameter, but if you think of sensitivity in terms of signal-to-noise (and hence limit of detection) then you could differentiate methods by either;

1. LOD = 3 x (noise)pp

here, LOD is initially determined as a function of peak height. In the alternative case we have a function of peak area:

2. LOD = 3.3 x sd / S

where,
(noise)pp = measured peak-to-peak noise in the baseline
sd = std dev of regression line
S = slope of regression line in calibration curve

In regulatory terms the first is just as acceptable, however you should go on to demonstrate your calculated value is suitable whichever you use.
2 posts Page 1 of 1

Who is online

In total there are 21 users online :: 1 registered, 0 hidden and 20 guests (based on users active over the past 5 minutes)
Most users ever online was 5108 on Wed Nov 05, 2025 8:51 pm

Users browsing this forum: Georguslat and 20 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry