Methods for monitoring system pressure over time

Discussions about HPLC, CE, TLC, SFC, and other "liquid phase" separation techniques.

5 posts Page 1 of 1
Hi all,

This is my first post to the forum.

We are currently looking for a good method for tracking pressure increases over time in order to monitor column lifespan. While this can be done with a chromatogram of the pressure reading, interpreting it is somewhat subjective. It would be nice to have a single number to use for trending.

We are using a Waters 2695 with Empower 2 software. It is an RP-HPLC method with a gradient elution.

My initial thought was to apply a mean smoothing to the pressure chromatogram, and then integrate the entire pressure trace (to account for the changing pressure during the gradient) as a single peak with a baseline of 0 psi. I have figured out how to force the integration as a single peak using the "Force Peak" event, but can't find a way to force a baseline of 0 psi.

If this is not possible, is there a way to extract a pressure value from a fixed time point (or even better, over a time range) and report this as a number?

Does this make sense? Any ideas? What do other folks do to track pressure increases between runs?

Thanks!

edit: I do have one semi-kludgy solution, which is to end the gradient with a short period of 0 flow rate, which should cause the pressure to drop to 0, and then force the baseline backward from that point. I'd prefer something a little cleaner though . . .

What is a pressure chromatogram? Normally one would attempt to clean a column when the pressure gets to an unacceptable level.

Although there are exceptions, we have found that logging pressure at the start of each run for a given column allows us to notice when back pressure has increased appreciably over time (to the extent that a particular column is used for a particular method).
Thanks,
DR
Image

You can collect a system pressure channel from you 2695 into Empower.

With that data you have a couple of choices.

1. The hard way. Manually look at each pressure trace and write down the system pressure at a constant time point (usually at the very beginning or very end of the run). Plot the data in a spreadsheet.

2. The easy way. Write a custom calculation within empower to do "the hard way" for you. Use a control chart report within empower to display the data.

Thanks for the replies!
You can collect a system pressure channel from you 2695 into Empower.

With that data you have a couple of choices.

1. The hard way. Manually look at each pressure trace and write down the system pressure at a constant time point (usually at the very beginning or very end of the run). Plot the data in a spreadsheet.
That's what I'm doing now. Manual examination of the pressure trace.
2. The easy way. Write a custom calculation within empower to do "the hard way" for you. Use a control chart report within empower to display the data.
This is more in-line with what I want to be doing. I'll look into the custom calculations and see what I can figure out (I'm pretty new to this software, hence my questions). Thanks for pointing me in the right direction.
5 posts Page 1 of 1

Who is online

In total there are 203 users online :: 4 registered, 0 hidden and 199 guests (based on users active over the past 5 minutes)
Most users ever online was 5108 on Wed Nov 05, 2025 8:51 pm

Users browsing this forum: Ahrefs [Bot], Amazon [Bot], Bing [Bot], Google [Bot] and 199 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry