by
cygg168 » Thu Feb 18, 2010 9:58 pm
Thanks every one.
I have tried everything I can think of. The tubing into the column and out to the detector has been changed 0.007 and 0.005 respectively, and as short as possible. One thing puzzles me is that we use the same instrument and detector (even the same electrode settings) but different mobile phase and column for another impurity test method of the same product, and there is no problem for that test at all.
The method requires to use a 50 cm guard column for a 150 cm analytical column. Without the guard column, the retention time is about 28 minute, but the plate count is only 800. With the guard column, the retention time is almost 50 minutes and can pass the plate count requirement but the peak shape is still broad. Increasing the column temperature from 22C to 25 will increase the retention time from 47 minutes to 67 minutes.
The system takes very long time to equilibrate each time when a condition changed. For example, when the column temperature changed from 22C to 25C, it took more than 6-8 hours for the retention time became consistant. Is this normal for the Amperometric electrochemical detector?
I will check the electronic troubleshooting. I have to consultant with the chemist who setup the instrument.