Posted: Tue Feb 16, 2010 5:12 pm
Hi,
I have changed the filament and the multiplier in the trap one month ago, the calibration and check air water were ok during all this time but the sensitivity was unstable. I could dectect all the compounds and I have a good base line, however when I select diagnostic I get the following message:
THE INTEGRATOR ZERO CAN NOT BE PROPERLY ADJUSTED (READS 0 ADC COUNTS WHEN SET TO MAXIMUM). LIKELY CAUSES INCLUDE:
1) EITHER R20 OR R66 ARE MISADJUSTED ON THE SAPWAVE BOARD OR 2) THE MANIFOLD ELECTRONICS ASSEMBLY IS DEFECTIVE (ELECTROMETER) OR 3) THE ELECTROMETER INPUT IS SHORTED OR 4) THE SAPWAVE BOARD IS DEFECTIVE (INTEGRATOR/LPF).
Basic Heater Test: Passed
Ionization Test: Passed
FAILED.
Any help will be very appreciated
Thanks a lot
I have changed the filament and the multiplier in the trap one month ago, the calibration and check air water were ok during all this time but the sensitivity was unstable. I could dectect all the compounds and I have a good base line, however when I select diagnostic I get the following message:
THE INTEGRATOR ZERO CAN NOT BE PROPERLY ADJUSTED (READS 0 ADC COUNTS WHEN SET TO MAXIMUM). LIKELY CAUSES INCLUDE:
1) EITHER R20 OR R66 ARE MISADJUSTED ON THE SAPWAVE BOARD OR 2) THE MANIFOLD ELECTRONICS ASSEMBLY IS DEFECTIVE (ELECTROMETER) OR 3) THE ELECTROMETER INPUT IS SHORTED OR 4) THE SAPWAVE BOARD IS DEFECTIVE (INTEGRATOR/LPF).
Basic Heater Test: Passed
Ionization Test: Passed
FAILED.
Any help will be very appreciated
Thanks a lot