Troubleshooting with GCMS Saturn 2000

Discussions about GC-MS, LC-MS, LC-FTIR, and other "coupled" analytical techniques.

4 posts Page 1 of 1
Hi,

I have changed the filament and the multiplier in the trap one month ago, the calibration and check air water were ok during all this time but the sensitivity was unstable. I could dectect all the compounds and I have a good base line, however when I select diagnostic I get the following message:
THE INTEGRATOR ZERO CAN NOT BE PROPERLY ADJUSTED (READS 0 ADC COUNTS WHEN SET TO MAXIMUM). LIKELY CAUSES INCLUDE:
1) EITHER R20 OR R66 ARE MISADJUSTED ON THE SAPWAVE BOARD OR 2) THE MANIFOLD ELECTRONICS ASSEMBLY IS DEFECTIVE (ELECTROMETER) OR 3) THE ELECTROMETER INPUT IS SHORTED OR 4) THE SAPWAVE BOARD IS DEFECTIVE (INTEGRATOR/LPF).
Basic Heater Test: Passed
Ionization Test: Passed
FAILED.

Any help will be very appreciated
Thanks a lot :D

I used to run a Varian 2000 but never run into this particular message. One hint, vent the MS and make sure there are no shorts to ground from any of your ion source components. I would thoroughly check the filament leads...using a mulitmeter once the source is removed and check nothing is physically touching to ground. Easily checked with a simple multimeter. Also make sure your electron multiplier is not touching the source block. This is a sliding mechanism that can be easily moved too far forward. Its designed to be removed easily so that you can replace the detector. Have you recently cleaning the trap? These are some of the simple things to check for a start. Timothy

Hi Sabikh,
I have had similar problems with Saturn. I needed a new SAPWAVE board in the end. I couldn't even get through the integrator zero part of the auto-calibration.
I would still try the trap cleaning procedure though first as recommended.
WK
I'm Sorry I Haven't A Clue - Just A Minute - The Unbelievable Truth
Hi,

Thank you so much for your advices, I found that I had a shortage between the filament and the trap, I am going to change the SAPWAVE .
Thanks a lot
HS
4 posts Page 1 of 1

Who is online

In total there are 54 users online :: 1 registered, 0 hidden and 53 guests (based on users active over the past 5 minutes)
Most users ever online was 5108 on Wed Nov 05, 2025 8:51 pm

Users browsing this forum: Bing [Bot] and 53 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry