Varian Saturn 2000 MS Heater Failure--Keeps burning out

Discussions about GC-MS, LC-MS, LC-FTIR, and other "coupled" analytical techniques.

4 posts Page 1 of 1
Hello,
The trap heater keeps burning out on a Saturn 2000 MS. I would appreicate if anyone could give me a tip for why, or what to look for.
Thanks,
FB
Hi all, it looks like I have a trap heater failure. Is this common? Is it easy to replace?
Just a thought, but there are two heaters for this instrument. One is rated for 120VAC and one for 220VAC. Make sure you have the correct P/N for the voltage you'running your instrument on. Hope this helps.
Hello,

I know, that Varian was a bad series of heater which burned out easily.
The part no. of the 240V heater assy is 393031392, but You can buy only the heater from the net and You need just fixing the cables. It is much cheaper!
It is not so complicate to change, but need to be watchfully when you take of the manifold boards.

If You are interesting only the cartridge, I will looking for You.

Regards,
4 posts Page 1 of 1

Who is online

In total there are 42 users online :: 2 registered, 0 hidden and 40 guests (based on users active over the past 5 minutes)
Most users ever online was 5108 on Wed Nov 05, 2025 8:51 pm

Users browsing this forum: Bing [Bot], Google [Bot] and 40 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry