We're doing an HPLC analysis for 7 substances (beta-carbolines), and I'm using a multi-channel detection
For all substances, I'm using a 5-point calibration curve with doubled concentrations. The concentrations and channels for each analyte are as below:
Harmol (247 & 320 nm), harmane, norharmane (247 & 300 nm): .125, .25, .5, 1, and 2 ppm
Harmine (320 nm): 10, 20, 40, 80, 160 ppm
Harmaline (374 nm): 31.25, 62.5, 125, 250, 500 ppm
THH (223 & 290 nm): 2.5, 5, 10, 20, 40 ppm
Harmalol (374 nm): 10, 20, 40, 80 ppm
At 320 and 373 nm, peak areas doubled at the first concentrations but decreased at the third point, then the area slightly increased at the fourth calibration point and doubled at the fifth point, especially for harmine and harmaline. I'm not observing a similar problem at the standards at 223 nm, 247 nm, 290 nm, or 300 nm.
My injection volume was 20 μL, and I decreased it to 10 μL, but I'm still encountering this problem.
I'm preparing the calibration curves by mixing all standards with serial dilution. Therefore, it should not be a miscalculation problem.
I cannot find the reason behind the problem. Could anyone encounter a similar problem?
Thank you in advance!