Column generating pressure

Discussions about HPLC, CE, TLC, SFC, and other "liquid phase" separation techniques.

3 posts Page 1 of 1
Hello everyone

I am currently working on carotenoids. I am using a 1200 series HPLC witha a pressure limit of 400 bar and a YMC carotenoid column (150 X 4.6mm S-5) to separate my carotenoids. I have a total of 300 samples to run.

For my extraction, I used acetone:chloroform (1:1).

The gradient I use is as follows:
mobile phase A: TBME
mobile phase B: methanol
Gradient: 0 min A is 15%, 0.5 A is 15%, 1.5 A is 35%, 3.0 A is 50%, 4.5 A is 80%, 9.0 A is 80%, 13 A is 15%.

The flow rate is 0.8ml/min

The injection volume is 10 microliters.

Now the problem

I first used an old column and discovered that whenever I completely run my 100 samples, the column generates a high pressure in the HPLC system, reaching about 250 bar and above.

Upon troubleshooting, I found that the pressure was coming from the column and not due to tube clogging or any other system-related issue.

Afterward, we decided to buy a new column, and after 100 samples, the same problem occurred. We experienced rapid pressure increase.

Using a flow rate of 0.8 ml/min, the ideal pressure is usually about 60-80 bar. However, after 100 samples run, there is a sharp rise in pressure generated from the column

Procedures to correct this problem

I carried out a reverse wash using the following reagent:
    Methanol and TBME (to remove polar compounds in the column)
      100% TBME (to remove polar compounds in the column)
        100% isopropanol
          Water coupled with increasing the temperature of the column to 50 degree.
            I Carried out overnight wash using TBME and methanol
            All above steps were done with a low flow rate of 0.3ml/min.

            While sometimes the column pressure drops, it only takes a while before it spikes again.

            Please, what can I do in this case, and what could be going wrong, as it is non-economical to keep purchasing new columns?
            In short, the problem you are experiencing is certainly due to what you inject in it. Things you may consider if not yet done:
            -centrifuge twice or filter to make sure the extracts are totally clear and particle-free.
            -add a guard column before you column (cheaper to replace than a column itself).
            -decrease the hydrophobicity of your extraction solvent so it becomes more compatible with the intital mobile phase composition.
            I would add to Gaetans post:

            Possibly you have some compounds precipitating from yourr aceton/CHCl3-extract when it comes in contact with the initial mobile phase that is much more polar.

            You could check this in a test tube by diluting some of your extract with the mobile phase.

            If you do not want to change your extraction procedure, you'll have to dilute your extract with methanol/TBME and filter. You also need to check how this influences the recovery of your analytes.
            3 posts Page 1 of 1

            Who is online

            In total there are 403 users online :: 0 registered, 0 hidden and 403 guests (based on users active over the past 5 minutes)
            Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am

            Users browsing this forum: No registered users and 403 guests

            Latest Blog Posts from Separation Science

            Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

            Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

            Liquid Chromatography

            Gas Chromatography

            Mass Spectrometry