Thank you for your reply, Tyler. I very much appreciate it. You wrote (reply between lines):
> Lots of questions left unanswered: what are your analytical conditions?
22 x 250 mm Vydac 218TP 15-20 µm C18 column
Flow rate = 25 ml/min A = 10 mM ammonium acetate, 8% acetonitrile, pH 5.0, B = acetonitrile, 0.5 ml injected in a 5 ml loop, ~40 mg of compound injected
> Is this a new phenomena or is this the first prep run for this analyte?
Not new. Here's a similar run with another very similar compound:
https://ibb.co/sv4kCZjW
And a run of the same compound but with more loaded (~100 mg in 4 ml), and on a different gradient (NOTE: gradients on the 650e system pump are in increments of 1% B, so the gradient is actually a step gradient as noted in the plots).
https://ibb.co/27Brjnt8
> How does the sample prep look for the analytical and prep sample?
I take the compound down to a crust by rotary evaporation and dissolve in in A, filter, inject.
> Could this be a result of an over-concentrated/loaded sample, what is the maximum concentration and volume you can inject and retain relatively normal peak-shape/chromatography?
I used to run the better part of half a gram of a mixture of several very similar compounds (natural product extract) on the same column using the same pump, and I got relatively normal peak shapes as shown below (some peaks maxed out, it was an analytical detector):
https://ibb.co/ymKS1ZnC
I don't think it's a capacity issue. I've also used the prep pump as pump for a 4.6 x 250 mm analytical column at 2 ml/min with basically no difference between it and a dedicated analytical system in terms of peak shape, though the baseline at 210 nm was bumpy. Which is why I bought the dedicated analytical system, and worry about pump seals and check valves. However, I checked the output of the pump, and the volumes are correct across the %B spectrum.
> These analytical runs that you have posted in particular, do these peaks have an identical RT or is this an adjusted image to make the peaks look nice and congruent?
They have RT within about 0.01 min, I'm quite sure they're effectively superimposable and represent the same compound with various levels of contaminant peaks. The plots are adjusted by the software so that the maximum height peak is of the same height in every plot, obviously from the drizzled peak plot the size of the peaks I clumped together is not to scale, but they are eluting at essentially the same time.
Happy to elaborate or answer other questions, I'd really like to get output from this prep column that resembles the peaks again.