Hello everyone,
This topic may be not very new but still confusing for me as an early-career chromatographer. I'm trying to hook Shimadzu ELSD to an Agilent HPLC, what I've figured so far is 1V Output will go to 3-pin Analog in port in the A/D box (we use Agilent UIB II).
My questions are where should the Ext. events and Ext. Autozero cables connect to? and
Should I connect the A/D box to the computer to collect the signal?
I'm retired, and not well versed in the "newer" Agilent instruments. But when we had external detectors, we used an A-D "box" and it was IMPORTANT to have two cables to connect it to computer: both GPIB and the 9-pin remote cable. Some folks forgot the remote cable...not good.
Agilent APG-remote cable has two pins crucial for system functioning.

APG-ready - all nodes try to DOWN (0) that signal to indicate readiness. And maintain it HIGH when non-ready.
APG-start - all nodes should start when this signal goes DOWN.
3 posts Page 1 of 1

Who is online

In total there are 2 users online :: 1 registered, 0 hidden and 1 guest (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am

Users browsing this forum: Semrush [Bot] and 1 guest

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry