Cal levels 5K, 4K, 2K, 1K 800, 400, 200, 120, 80, 50, 40, 20, 10 ppt
Response 136768 for 5000 ppt
Response 24320 for 1000 ppt
Response 8969 for 400 ppt
Response 1152 for 50 ppt
Response 487 for 20 ppt
Response 181 for 10 ppt
I am using 400 ppt as my CCV. Fluorobenzene and Chlorobenzene-d5 at 5 ppm as my internal standards. I tried 1 ppm but could not get consistent passing of BFB tune. 5 ppm passes without a problem. Fluorobenzene response is 650,000 and Chlorobenzene-d5 response 2,000,000 and 4-BFB response is 200,000.
I figure with my 5mL sample and a 40:1 split inlet that purge and trap puts ~5 times less 1,2-EDB on the detector than method 8011. But then ECD may also be much more senstive than an MSD.
Any comparisons would be welcome.