I have noticed for some time that we have been getting some evidence of analyte peaks in blank samples. I have done all the usual troubleshooting things and have come to the conclusion that an errant setting during some large volume injection optimisation a while back has most likely led to backflash causing our carryover. Since then I have of course changed all relevant consumables several times over and have cleaned all parts of the inlet that I can access suggesting that the contamination is most likely sitting in one of the lines.
During a recent visit by an engineer I discussed this and possible solutions and his viewpoint was that although it is possible to change out or clean the split vent line, because the system (Agilent 8890) has a PAL3 unit sitting on top, it's too difficult and time consuming to do and suggested 'injecting 10ul of hexane' with the 'detector [MSD] switched off'. At the time, I nodded sagely and we moved on to other issues that were more pressing, but realise in hindsight that I have no idea what he was actually suggesting I do...
Any suggestions as to what his intention was?
Kind Regards
TD2